My doctor helped publish this study recently and thought I'd post it for discussion.
https://pubmed.ncbi.nlm.nih.gov/40197751/
An exerp from the study:
Results: We identified a total of 410 cases: 220 BSCI and 190 CP devices. One hundred twenty-nine were revisions. Seventy-two met the inclusion criteria (63 BSCI and 9 CP). BSCI mechanical failures included: cylinder rupture, 26/63 (41.3%), tubing fracture, 7/63 (11.1%), reservoir rupture, 3/63 (4.8%), cylinder aneurysm, 6/63 (9.5%), and pump failure, 21/63 (33.3%). Coloplast mechanical failure included: tubing fracture 7/9 (77.8%), while reservoir rupture and cylinder aneurysm each were 1/9 (11.1%). Time to mechanical failure was a median of 48 and 41 months, respectively, for BSCI and CP devices.
My doctor told me at my last checkup that Coloplast is planning to change their tubing material to Bioflex, the same material as their cylinders, to reduce their failures. Can anyone else confirm this?
Boston Scientific and Coloplast IPP Failure Study
-
Wooody
- Posts: 181
- Joined: Mon Jan 27, 2025 3:04 pm
- Location: Bay Area, CA
Boston Scientific and Coloplast IPP Failure Study
Titan Classic 22cm + 1cm RTEs - 2/25 - Dr Karpman, Bay Area CA
-
newhope
- Posts: 256
- Joined: Sun Dec 24, 2023 11:20 am
Re: Boston Scientific and Coloplast IPP Failure Study
a median of 48 months is pretty low. It means there are 50% of devices that fails within 4 years
LGX 18 + 1. 2023/12
Pre-OP VED length: 5.5". Post-OP: 4". At 3 months post-op: 5". At 1 yr: 5,3"
Cycling with a lot of pain for the first 5 months, now improving
If you have a motorcycle stop using it (ED with motorcycle accident).
Pre-OP VED length: 5.5". Post-OP: 4". At 3 months post-op: 5". At 1 yr: 5,3"
Cycling with a lot of pain for the first 5 months, now improving
If you have a motorcycle stop using it (ED with motorcycle accident).
-
Courage
- Posts: 201
- Joined: Thu Oct 26, 2023 8:51 pm
Re: Boston Scientific and Coloplast IPP Failure Study
newhope wrote:a median of 48 months is pretty low. It means there are 50% of devices that fails within 4 years
I assume that's the median for devices that failed within the study period, not the median for all devices installed.
Middle-aged SGM with lifelong ED. AMS 700 CX 21cm + 3.5cm RTEs implanted January 2025 and explanted due to infection February 2025, with salvage. Revision to Coloplast Titan 24cm + 1cm RTE July 2025.
-
ElbowRoom
- Posts: 636
- Joined: Mon Mar 17, 2025 1:58 pm
Re: Boston Scientific and Coloplast IPP Failure Study
Wooody wrote:My doctor told me at my last checkup that Coloplast is planning to change their tubing material to Bioflex, the same material as their cylinders, to reduce their failures. Can anyone else confirm this?
Dr. Hakky told me they are updating the tubing to increase durability, but with FDA approvals and such it was still 3-5 years out.
58yo Coloplast Titan 28cm Penoscrotal with Dr. Hakky 10/21/2025.
Pre-op erect measurements:
8.5"L and 6.5"C
Post-op: 8”L and 6”C at one week.
Pre-op erect measurements:
8.5"L and 6.5"C
Post-op: 8”L and 6”C at one week.
-
ragingbull
- Posts: 167
- Joined: Fri Jul 26, 2024 10:24 pm
Re: Boston Scientific and Coloplast IPP Failure Study
tooyoung wrote:Courage wrote:newhope wrote:a median of 48 months is pretty low. It means there are 50% of devices that fails within 4 years
I assume that's the median for devices that failed within the study period, not the median for all devices installed.
Yeah they cherry picked these faulty devices...![]()
"A sample is not a sample"...beautiful....whatever makes you sleep at night.
Come on! Let's keep a positive attitude, take every step in life like you own that sidewalk, hold your head high, shoulders back! 50% of these devices are still surviving after 20 years in assisted nursing homes where they're used for non-stop masturbation sessions that'd put porn stars to shame, blush and beg for a timeout!
72, Married for 30 years to a beautiful nympho. 2025 Titan Classic L 22, G 6 after multiple surgeries. Best gift of my life.
-
wanttokno
- Posts: 130
- Joined: Tue Jun 11, 2013 9:33 am
- Location: Delaware
Re: Boston Scientific and Coloplast IPP Failure Study
I’ve got to say that looking at my third revision in just eleven years I’m finding this information a bit disheartening but at this point there’s no other real option. Oh well, you do the best that you can with what you’ve got.
Thanks for the post.
Thanks for the post.
62 yrs old
Pre implant size 7.8 in L x 6 in G
Implanted 04/14 w/ an LGX 18cm +3
Size 7.6-7.7 in L x 5 G
65 ml res w/45 ml
Revision 04/20
Titan 22 cm Size 7.5 in L x 5.5 G
125 ml res
Revision 11/22
Titan 22 cm + 1 Size 7.5 in L x 5.5 G
125 ml res filled
Pre implant size 7.8 in L x 6 in G
Implanted 04/14 w/ an LGX 18cm +3
Size 7.6-7.7 in L x 5 G
65 ml res w/45 ml
Revision 04/20
Titan 22 cm Size 7.5 in L x 5.5 G
125 ml res
Revision 11/22
Titan 22 cm + 1 Size 7.5 in L x 5.5 G
125 ml res filled
-
Wooody
- Posts: 181
- Joined: Mon Jan 27, 2025 3:04 pm
- Location: Bay Area, CA
Re: Boston Scientific and Coloplast IPP Failure Study
I think we all share the goal of having better reliability of IPP products. As such, I hope that Coloplast does indeed improve their tubing to match the reliability of their cylinders and pumps as this seems to be the weakness of their IPP. Boston Scientific seems to need improvement in more of their components based on this study. Not sure what/if they have any plans for product improvements. Has anyone heard if they are going to make any product changes?
As far as this study goes, it does make sense to exclude surgeries that were the result of infections or other non-product failures such as reservoir migration, mis-sizing etc and only including actual product defects and failures.
As far as this study goes, it does make sense to exclude surgeries that were the result of infections or other non-product failures such as reservoir migration, mis-sizing etc and only including actual product defects and failures.
Last edited by Wooody on Wed Sep 03, 2025 2:30 pm, edited 1 time in total.
Titan Classic 22cm + 1cm RTEs - 2/25 - Dr Karpman, Bay Area CA
-
Chris1967
- Posts: 30
- Joined: Sun Jun 01, 2025 12:14 pm
Re: Boston Scientific and Coloplast IPP Failure Study
Hello guys,
If you have a good doctor and treat your implant normally, I think longer periods are feasible. I compare it to a car. If you drive until it's worn out, you'll need repairs sooner than if you drive moderately. If the implant is constantly pumped over 100 percent and used for rough anal sex, it will wear out faster than if you use it normally 3 to 4 times a week for sex. What do you think?
If you have a good doctor and treat your implant normally, I think longer periods are feasible. I compare it to a car. If you drive until it's worn out, you'll need repairs sooner than if you drive moderately. If the implant is constantly pumped over 100 percent and used for rough anal sex, it will wear out faster than if you use it normally 3 to 4 times a week for sex. What do you think?
57 years old, living in Germany, ED since 2004, Levitra in high doses still works.
I sometimes use an Osbon pump and two D-rings during sex.
I sometimes use an Osbon pump and two D-rings during sex.
-
newhope
- Posts: 256
- Joined: Sun Dec 24, 2023 11:20 am
Re: Boston Scientific and Coloplast IPP Failure Study
these devices are just hard plastic tubing, they should be bullet proof. The fact that they break within few years imho is a shame
These companies want to make more profit on the behalf of very vulnerable men that can't do anything else. And each revisions increases chances of infections and other terrible outcomes
There should be a class action against these companies
These companies want to make more profit on the behalf of very vulnerable men that can't do anything else. And each revisions increases chances of infections and other terrible outcomes
There should be a class action against these companies
LGX 18 + 1. 2023/12
Pre-OP VED length: 5.5". Post-OP: 4". At 3 months post-op: 5". At 1 yr: 5,3"
Cycling with a lot of pain for the first 5 months, now improving
If you have a motorcycle stop using it (ED with motorcycle accident).
Pre-OP VED length: 5.5". Post-OP: 4". At 3 months post-op: 5". At 1 yr: 5,3"
Cycling with a lot of pain for the first 5 months, now improving
If you have a motorcycle stop using it (ED with motorcycle accident).
-
Chris1967
- Posts: 30
- Joined: Sun Jun 01, 2025 12:14 pm
Re: Boston Scientific and Coloplast IPP Failure Study
Analysis
Long-Term Survival Rates of Inflatable Penile Prostheses: Systematic Review and Meta-Analysis
Larry E Miller et al. Urology. 2022 Aug.
Free article
Show details
Full text links
Cite
Abstract
The purpose of this study was to determine long-term survival of inflatable penile prosthesis (PP) and identify potential factors that may influence device survival. We performed a systematic review of Medline, Embase, and the Cochrane Central Register of Controlled Trials for studies of men treated with inflatable PP with at least 5 years of device survival data. We performed a random effects meta-analysis to estimate device survival at 1, 3, 5, 10, 15, and 20 years of follow-up. The robustness of the meta-analysis results was evaluated in a 1-study removed sensitivity analysis and sources of heterogeneity among studies were investigated with subgroup analysis. In 12 studies (20,161 patients; median age 57 years), PP device survival was 93.3% at 1 year, 91.0% at 3 years, 87.2% at 5 years, 76.8% at 10 years, 63.7% at 15 years, and 52.9% at 20 years. The results of the meta-analysis were not significantly influenced by single study effects in a 1-study removed sensitivity analysis. In a subgroup analysis, 5-year device survival rates were 90.6% vs 82.1% (P = .01) comparing newer vs older studies; no other patient or study design characteristic was statistically associated with device survival rates. In conclusion, the median device survival time of an inflatable PP is approximately 20 years. The factors responsible for improved device survival in newer studies remain unclear and warrant further study.
Copyright © 2022 The Author(s). Published by Elsevier Inc. All rights reserved.
PubMed Disclaimer
Long-Term Survival Rates of Inflatable Penile Prostheses: Systematic Review and Meta-Analysis
Larry E Miller et al. Urology. 2022 Aug.
Free article
Show details
Full text links
Cite
Abstract
The purpose of this study was to determine long-term survival of inflatable penile prosthesis (PP) and identify potential factors that may influence device survival. We performed a systematic review of Medline, Embase, and the Cochrane Central Register of Controlled Trials for studies of men treated with inflatable PP with at least 5 years of device survival data. We performed a random effects meta-analysis to estimate device survival at 1, 3, 5, 10, 15, and 20 years of follow-up. The robustness of the meta-analysis results was evaluated in a 1-study removed sensitivity analysis and sources of heterogeneity among studies were investigated with subgroup analysis. In 12 studies (20,161 patients; median age 57 years), PP device survival was 93.3% at 1 year, 91.0% at 3 years, 87.2% at 5 years, 76.8% at 10 years, 63.7% at 15 years, and 52.9% at 20 years. The results of the meta-analysis were not significantly influenced by single study effects in a 1-study removed sensitivity analysis. In a subgroup analysis, 5-year device survival rates were 90.6% vs 82.1% (P = .01) comparing newer vs older studies; no other patient or study design characteristic was statistically associated with device survival rates. In conclusion, the median device survival time of an inflatable PP is approximately 20 years. The factors responsible for improved device survival in newer studies remain unclear and warrant further study.
Copyright © 2022 The Author(s). Published by Elsevier Inc. All rights reserved.
PubMed Disclaimer
57 years old, living in Germany, ED since 2004, Levitra in high doses still works.
I sometimes use an Osbon pump and two D-rings during sex.
I sometimes use an Osbon pump and two D-rings during sex.
Who is online
Users browsing this forum: AhrefsBot, body2body, brickman, Chriss, ClaudeBot, lornezm and 32 guests
